EH&S wants you to get out of your lab safely in an emergency. A good question to ask yourself is: “Could I quickly and safely get out of my lab in the dark?” Maintain a minimum of 36 inches clearance in aisles at all times and never block required exits with equipment, boxes, or other items (required exits have a lighted “exit” sign above the door).
Flammable gases (e.g., hydrogen, methane, propane, acetylene) and oxygen (or other oxidizing gases) must be separated by 20 feet or by a non-combustible partition extending 18 inches above and to the sides of stored gas cylinders. Cylinders should be positioned so that the label with the name of the gas and its hazard is clearly visible. Gas cylinders are often difficult to reach; in an emergency, first responders need to see what they’re dealing with as quickly as possible.
For fire sprinklers to work properly, they must not be blocked in any way such that their spray pattern is impeded. There needs to be enough room for the spray from fire hoses to reach throughout the room. Any storage within 18” of sprinklered areas or 24” of non-sprinklered areas of the ceiling must remain clear, except when the storage is along a wall.